Topic: Cyanide analysis - Samples storaging time  (Read 6111 times)

0 Members and 1 Guest are viewing this topic.

Offline tripton

  • Regular Member
  • ***
  • Posts: 42
  • Mole Snacks: +0/-0
Cyanide analysis - Samples storaging time
« on: March 13, 2008, 10:46:05 AM »
Hi, I must analyze total CN in wastewaters. I will receive several samples and I haven´t distillation enough equipments to treat them in the reception day. The expected concentration is between 1 and 200 ppm.
In the Standard Methods for Examination of water & wastewater 20th ed. (APHA, AWWA, WEF) (page 1-33) I found the "recommended storaging time" is 24hs  :( but the "regulatory" is 14days  :).
Someone has experience analyzing this kind of analysis? Can I wait 2 or 3 days before to do the distillation?

Thanks in advance,

Tripton

Offline AWK

  • Retired Staff
  • Sr. Member
  • *
  • Posts: 7976
  • Mole Snacks: +555/-93
  • Gender: Male
Re: Cyanide analysis - Samples storaging time
« Reply #1 on: March 13, 2008, 11:39:32 AM »
Why not? The regulatory time is 14 days
AWK

Offline tripton

  • Regular Member
  • ***
  • Posts: 42
  • Mole Snacks: +0/-0
Re: Cyanide analysis - Samples storaging time
« Reply #2 on: March 14, 2008, 08:07:16 AM »
Yes, but recomended is 24hs.
Really I´m not sure if after 24hs the cyanide in the samples may suffer some degradation.
I guest the regulatory is 14d because in a real wastewater taked from an industry, it should acceptable if after 14days the CN is under regulation limits. Perhaps because the regulation limits are very stringents contemplating this long analysis time.
In my case, this samples cames from an experiment. Then if I analyze some of them in different days and the Cyanide disappears, I may get wrong conclusions.
What´s your opinion about?

Thanks

Offline JGK

  • Chemist
  • Full Member
  • *
  • Posts: 738
  • Mole Snacks: +66/-19
  • Gender: Male
Re: Cyanide analysis - Samples storaging time
« Reply #3 on: March 17, 2008, 02:21:47 PM »
Why not test thestability yourself, spike water samples, test a sub-sample immediatelyand store the remainder. Then test at specified storage intervals and compare to the original value.

If you assign an acceptance criterial to the test, you can define your own storage period and have the data to back it up.
Experience is something you don't get until just after you need it.