I have been working with and doing research in journal articles and books to find a method, or step by step procedure to use in determining As and Se concentration by hydride generation. So far, I have found some useful infomation, but not enough to get this method to work for me.
I am using an MHS 15 in conjunction with an AAnalyst 400 spectrometer. I am currently working with As.
I am trying both 2% HCl and 2%HNO3 to make my blank and calibration standards. I am also reducing these by 2 methods:
group 1: reduction by adding 2 mL L-cysteine per 10 mL sample/solution.
group 2: reduction of As by adding 1 mL KI + L(+)-ascorbic acid per 10 mL sample/solution
Wait 30 minutes for reduction.
While I am waiting, I make up a fresh batch of 3% NaBH4 in 1%NaOH sol'n to be used as a reductant.
15 minutes prior to running, I heat the quartz cell.
I add 10mL 1.5% HCl diluent to the reaction flask and add 10uL of the 1ppm calibration standard which should give reading in the ppb range.
I wait the 20 second prereaction purge time and depress the plunger on the MHS 15 while the method for As is up and running on the AA400.
All recommended conditions are used for wavelength, etc, but I am getting no readings whatsoever.
I am having a lot of trouble with this, and the manual is vague. I also do Hg determinations using the same (but separate) machine (so it is not contaminated, it is new... I thought that might be the problem because SnCl2 is used for that). The recommended conditions for Hg determinations is not spot on either. There have been several modifications to make it work right, so I am thinking there are several modifications that must be done in order for As and Se determinations to work as well.
Does anyone have a method developed for this technique? Any suggestions on what I could try or what I am doing wrong? I have not been able to find the answers in any book, journal article, user's maunal, or through support tech. If anyone has a developed method, please *delete me*
Thanks!