Without a complete description, I cannot be certain, but if I understand you correctly, the answer is no. Since your samples have already been subject to environmental oxidation, I would rule out a rapid and complete oxidation to a new product. Therefore, the question is whether partial oxidation might have occurred. You did not describe that in your scenario. A partial oxidation should retain at least some of your original peaks and now should include new additional peaks. However, you describe a shift in retention time. That sounds like a change in column, solvent, flow rate, etc.