sulfide by iodometry confusion
« on: October 12, 2010, 02:46:52 AM »
in our lab we measure sulfides in water spectrophotometry or by the method known as methylene blue method , but due to the highly interfering ions present in the water we intended to use the other method iodometry titration.

after prepering the sample and performing titrations the volume of thiosulfate used was very high so that according to calculations the concentration of sulfides was in minus sign ....lol!!!!

i dont know what can i do ?!!!!1
i supposed the following :

- may thiosulfate react with other component in the sample together with iodine.
- may be interfering ions still present (so i washed the sample many times - still the confusion present)

i dont know what is the problem can any body help????

Offline el13

  • Regular Member
  • ***
  • Posts: 16
  • Mole Snacks: +0/-1
Re: sulfide by iodometry confusion
« Reply #1 on: October 12, 2010, 05:35:47 AM »
you can try potentiometric method, using AgNO3 solution.

Offline aeacfm

  • Full Member
  • ****
  • Posts: 226
  • Mole Snacks: +7/-2
  • Gender: Male
Re: sulfide by iodometry confusion
« Reply #2 on: October 12, 2010, 06:31:48 AM »
we exactly compare between many methods to finally use precise and fast method .

but still want to know what happened there ?

Offline Pradeep

  • Regular Member
  • ***
  • Posts: 95
  • Mole Snacks: +5/-3
Re: sulfide by iodometry confusion
« Reply #3 on: October 12, 2010, 12:27:46 PM »
Sulfide is at its minimum oxidation  number (-2). Therefore liberation of iodine is not possible according to me. Did you got a blue color with starch ? But at higher acidity thiosulphate can be reacted by depositing sulfur. 

Offline aeacfm

  • Full Member
  • ****
  • Posts: 226
  • Mole Snacks: +7/-2
  • Gender: Male
Re: sulfide by iodometry confusion
« Reply #4 on: November 09, 2010, 09:25:13 AM »
Sulfide is at its minimum oxidation  number (-2). Therefore liberation of iodine is not possible according to me. Did you got a blue color with starch ? But at higher acidity thiosulphate can be reacted by depositing sulfur.  

i cant understand you
could you please explain more ?

Offline AWK

  • Retired Staff
  • Sr. Member
  • *
  • Posts: 7978
  • Mole Snacks: +555/-93
  • Gender: Male
Re: sulfide by iodometry confusion
« Reply #5 on: November 09, 2010, 10:26:29 AM »
Are you sure you did calculations correctly. Show them!
AWK

Offline aeacfm

  • Full Member
  • ****
  • Posts: 226
  • Mole Snacks: +7/-2
  • Gender: Male
Re: sulfide by iodometry confusion
« Reply #6 on: November 10, 2010, 03:52:24 AM »
Are you sure you did calculations correctly. Show them!

I2 = 0.0187 N
Na2S2O3 = 0.0246 N

100 ml sample  + 25 ml I2 soln took  55ml Na2S2O3....

so equivalents of S-- in sample would be = (0.0187 N *25 ml ) - (0.0246 N *55 ml )
which give us minus no of equivalents = -0.8855 eqs

Offline AWK

  • Retired Staff
  • Sr. Member
  • *
  • Posts: 7978
  • Mole Snacks: +555/-93
  • Gender: Male
Re: sulfide by iodometry confusion
« Reply #7 on: November 10, 2010, 04:14:09 AM »
Error in concentration of iodide is the most probable, eg 0.187
AWK

Offline aeacfm

  • Full Member
  • ****
  • Posts: 226
  • Mole Snacks: +7/-2
  • Gender: Male
Re: sulfide by iodometry confusion
« Reply #8 on: November 10, 2010, 04:27:02 AM »
Error in concentration of iodide is the most probable, eg 0.187

 i dont think , because i was going to prepare 0.025 N but mis-weighting make it 0.0187
any way thanks alot

Offline AWK

  • Retired Staff
  • Sr. Member
  • *
  • Posts: 7978
  • Mole Snacks: +555/-93
  • Gender: Male
Re: sulfide by iodometry confusion
« Reply #9 on: November 10, 2010, 08:19:20 AM »
Check your all calculation when preparing solutions
AWK

Offline aeacfm

  • Full Member
  • ****
  • Posts: 226
  • Mole Snacks: +7/-2
  • Gender: Male
Re: sulfide by iodometry confusion
« Reply #10 on: November 10, 2010, 08:43:37 AM »
i can understand that , but why you exclude other factors?

Offline AWK

  • Retired Staff
  • Sr. Member
  • *
  • Posts: 7978
  • Mole Snacks: +555/-93
  • Gender: Male
Re: sulfide by iodometry confusion
« Reply #11 on: November 10, 2010, 10:11:44 AM »
If  your sample, in fact, contains sulfides, an  error in preparing solutions is the only possibility.
AWK

Offline aeacfm

  • Full Member
  • ****
  • Posts: 226
  • Mole Snacks: +7/-2
  • Gender: Male
Re: sulfide by iodometry confusion
« Reply #12 on: November 10, 2010, 10:17:45 AM »
even if sulfides traces ?
no possibility for interferene ?