Very much appreciate your help Corribus! Thank you also for the links. Should have thought of Wikipedia.
I made a blank and ran seven analysis. I also removed the cuvette and replaced as I find abs values will change appreciably compared to leaving in place and since all the samples are placed in the holder manually, thought this would give a better representation for LOD.
I then took the resulting abs value and converted to ppb concentration based on the calibration curve. Then I took the average ppb of the seven analysis of the blank and added this to 3.3 times the standard deviation. This method gave a LOD of 3.9ppb and a LOQ of 9.5ppb. Those values I have far better confidence in believing.
Did I do this correctly?
Concentrations I got from abs values of the blank in ppb were:
-0.565, 1.01, 1.644, 1.807, 1.641, 1.653, 1.023.
Avg blank conc: 1.17 ppb
stdev: 0.83