Topic: LOD Calculation  (Read 1972 times)

0 Members and 1 Guest are viewing this topic.

Offline Hend Yamani

  • Very New Member
  • *
  • Posts: 2
  • Mole Snacks: +0/-0
LOD Calculation
« on: June 15, 2015, 09:43:46 AM »
Can anyone please provide me with a reference that use calibration curve other than the original calibration curve of the analyte to calculate LOD?
For example; if the original calibration curve from (1-50 µg/mL), another calibration curve from (0.1-1 µg/mL) is constructed to calculate LOD.

Offline Hunter2

  • Sr. Member
  • *****
  • Posts: 2313
  • Mole Snacks: +191/-50
  • Gender: Male
  • Vena Lausa moris pax drux bis totis
Re: LOD Calculation
« Reply #1 on: June 15, 2015, 09:50:41 AM »
I don't know really what you mean. I guess LOD = Limit of detection. And the range is 1-50 µg/l then its not possible to detect below 1 µg/l because of the performance of the analytical method, machine, chemicals, etc. So its not possible to get  reasonable results with a other curve what is constructed below. 0.1-1 µg/l

Offline Arkcon

  • Retired Staff
  • Sr. Member
  • *
  • Posts: 7367
  • Mole Snacks: +533/-147
Re: LOD Calculation
« Reply #2 on: June 15, 2015, 11:25:52 AM »
Generally, LOD is defined as 3 times the noise for a representative background interval.  If you want to use a different calibration curve other than the original, I'm guessing you mean to perform a new assay.  Then, you have to re-develop it according to what your noise is.