December 22, 2024, 05:26:41 PM
Forum Rules: Read This Before Posting


Topic: UV Visible Spectroscopy calculation: Beer-Lambart's law  (Read 2736 times)

0 Members and 1 Guest are viewing this topic.

Offline STM

  • Regular Member
  • ***
  • Posts: 45
  • Mole Snacks: +0/-0
UV Visible Spectroscopy calculation: Beer-Lambart's law
« on: June 21, 2015, 11:55:55 AM »
Dear all,
Please I need clarification on this calculation procedure.

I did a UV-Visible measurement of Conjugated diene in lipid by measuring absorbance at 234 nm. I got an Absorbance = 3.31. Using molar absorbtivity, e, = 24000 M-1 cm-1, and path length = 1 cm.

From the formula A= eCL, C=A/eL = 3.310/ 24000x1 =0.0001379 M (137.9uM).

The weight of lipid (Oil) sample taken was 0.018g and was dissolved in 3.5 ml methanol as medium for the UV-Vis measurement.

I am expected to express the answer as umole/g (micromole/g). Is it just the 137.9 uM/0.018g or will there be need for other parameters to come into the calculation.

Please I need you advice on how to get to umole/ g. I have done some calculations already but I just need some clarifications from you all.

Thank you.

Offline Borek

  • Mr. pH
  • Administrator
  • Deity Member
  • *
  • Posts: 27885
  • Mole Snacks: +1815/-412
  • Gender: Male
  • I am known to be occasionally wrong.
    • Chembuddy
Re: UV Visible Spectroscopy calculation: Beer-Lambart's law
« Reply #1 on: June 21, 2015, 01:50:04 PM »
137.9 μM is the concentration of the diene in the solution, not its amount in the original sample.
« Last Edit: June 21, 2015, 03:30:28 PM by Borek »
ChemBuddy chemical calculators - stoichiometry, pH, concentration, buffer preparation, titrations.info

Offline Babcock_Hall

  • Chemist
  • Sr. Member
  • *
  • Posts: 5716
  • Mole Snacks: +331/-24
Re: UV Visible Spectroscopy calculation: Beer-Lambart's law
« Reply #2 on: June 22, 2015, 09:53:52 AM »
Maybe I am missing something here and I am not an analytical chemist, but an absorbance of 3.31 is out of the linear range of most spectrophotometers, due to stray light.  If stray light is a problem, there is an apparent negative deviation from the Beer-Lambert law.

Offline marquis

  • Full Member
  • ****
  • Posts: 433
  • Mole Snacks: +38/-3
Re: UV Visible Spectroscopy calculation: Beer-Lambart's law
« Reply #3 on: June 22, 2015, 10:29:40 PM »
Babcock Hall has a good point.

Typically, with Beer-Larmbert, linearity has to be proven.  It
can't be assumed.

If you want a quick test, do a 1:10 dilution and rerun the test.
See if the answer comes in at one tenth the value.  If it does, all is
good with the world.

Offline STM

  • Regular Member
  • ***
  • Posts: 45
  • Mole Snacks: +0/-0
Re: UV Visible Spectroscopy calculation: Beer-Lambart's law
« Reply #4 on: June 23, 2015, 12:28:02 PM »
Thank you all for your response. I am grateful. The points on the linearity range for Beer-Lamberts law is noted. Although, I do not intend to use the measurement for any form of calibration. Nevertheless, the result of the 10-fold dillution was still consistent.

Also, I will like to check this conversion>

If an oil sample has peroxide value of 5 milliequivalent of Oxygen per Kg, density = 0.889 g/ml. Its conversion to mmol/L is?

If 1 mmol of O2 = 2 meq, therefore, 5 meq = 2.5 mmol

5 meq/kg  x 1/2 = 2.5 mmol/kg.       2.5 mmol/Kg x 1 kg/1000 g x 0.889 g/ml x 1000 ml/L = 2.22 mmol/ L (2.22 mM)

Can you please confirm this conversion?

Thank you

Sponsored Links