The causes of pressure changes in your HPLC equipment could be from mechanics and/or chemical origin.
If the Pressure changes, check the flow rate indirectly with measurement Dead Time (tm), in this way you can pinpoint the cause: FLOW or CHEMISTRY. Also you must check LEAKS.
If you have permanent small fluctations you must check:
+ Air bubles (in the pump, in pump head seal defect , frit of suction is clogged)
+ Pump head seal defect
If your pressure increases, deacreses, increases you must check:
+ Inlet valve is clogged (you must check pump head malfunction)
NOTE: Check in your system defect on pressure sensor
Regards,
Chiralic