December 27, 2024, 01:50:41 PM
Forum Rules: Read This Before Posting


Topic: Uv-Vis Spectrophotometry Gaps in spectra  (Read 3299 times)

0 Members and 1 Guest are viewing this topic.

Offline 08roselx

  • New Member
  • **
  • Posts: 8
  • Mole Snacks: +0/-0
Uv-Vis Spectrophotometry Gaps in spectra
« on: December 02, 2019, 12:06:31 PM »
Hi

I am trying to calculate the SPF of sunscreens using the Mansur equation. I am using a method directly from a paper.

The solutions are made up in ethanol and the UV-Vis is blanked with the ethanol prior to analysis. 

I weigh 0.25g sunscreen and make up to 50ml then take 1ml and make to 25ml.

I took a spectrum of one formulation (made in triplicate) and I get breaks in the spectra where there is no absorption value?

I diluted the sample further as at some wavelengths I was getting absorption values of 4 and still the same issue. I also got negative absorption values. 

Can anyone help as to what this may be as I have never seen this before?
« Last Edit: December 02, 2019, 12:28:41 PM by 08roselx »

Offline Corribus

  • Chemist
  • Sr. Member
  • *
  • Posts: 3551
  • Mole Snacks: +546/-23
  • Gender: Male
  • A lover of spectroscopy and chocolate.
Re: Uv-Vis Spectrophotometry Gaps in spectra
« Reply #1 on: December 02, 2019, 01:50:55 PM »
Can you show a sample spectrum? An OD of 4 is way too high.  You should try to be below 1 if possible.
What men are poets who can speak of Jupiter if he were like a man, but if he is an immense spinning sphere of methane and ammonia must be silent?  - Richard P. Feynman

Offline wildfyr

  • Global Moderator
  • Sr. Member
  • ***
  • Posts: 1777
  • Mole Snacks: +203/-10
Re: Uv-Vis Spectrophotometry Gaps in spectra
« Reply #2 on: December 04, 2019, 10:35:15 AM »
I think your detector might be saturated because the concentration is way too high. You mean Absorbance values, not OD right?

I tend not to trust absorbance values over 2.

Offline 08roselx

  • New Member
  • **
  • Posts: 8
  • Mole Snacks: +0/-0
Re: Uv-Vis Spectrophotometry Gaps in spectra
« Reply #3 on: December 04, 2019, 05:04:35 PM »
The solvent that I had been using was contaminated which is why further dilution didn't help me. Ran samples again today using a different batch of solvent and all was fine :) I knew the absorbance was too high that's why I diluted the samples further (even though I knew it shouldn't have been high as I was following a standard protocol) but dilution didn't reduce the abs values as it was contaminated with some other substance.

Sponsored Links