Fluorimetry
« on: March 28, 2009, 06:09:35 AM »
Q.  The fluorescence intensity of a 0.65 ppm solution of riboflavin is 48.1. What is the concentration of riboflavin in a solution, when the fluorescence intensity for which is 57.8.  (Under the same instrumental conditions.)

 
-----
Usually we use standards to plot analysis curve to calculate the unknown concentration - However, in this problem only one set of data, and unknown concentration with Emission intensity of 57.8 is provided. How do you tackle with that?
 ???

If you know a trick or two, please let me know.


Offline Borek

  • Mr. pH
  • Administrator
  • Deity Member
  • *
  • Posts: 27887
  • Mole Snacks: +1816/-412
  • Gender: Male
  • I am known to be occasionally wrong.
    • Chembuddy
Re: Fluorimetry
« Reply #1 on: March 28, 2009, 06:39:57 AM »
Most likely you are to assume 0 for 0.