December 22, 2024, 09:06:51 PM
Forum Rules: Read This Before Posting


Topic: GC method using RRF instead of all compounds direct quantitation  (Read 3143 times)

0 Members and 1 Guest are viewing this topic.

Offline mariosm

  • New Member
  • **
  • Posts: 3
  • Mole Snacks: +0/-0
GC method using RRF instead of all compounds direct quantitation
« on: September 23, 2010, 11:41:07 AM »
Hi everyone,
I'm looking for a new method development approach or unconventional for GC quantitation method. We have 9 compounds to analyse simultaneously and I'm wondering if RRF (slope A/slope B) can be used for indirect quantitation? The goal would be to light the working standard preparation. For example, we would use an ethanol standard to quantify a mixture of ethanol, acetone and acetonitrile by direct quantitation for ethanol but indirect quantitation for acetone and acetonitrile using their respective RRF. Method would be validated using both calculation technics to prove equivalency but I'm wondering if this uncommon approach was seen in the past and if you have seen reference for this approach.
Regards

Offline Stepan

  • Chemist
  • Full Member
  • *
  • Posts: 358
  • Mole Snacks: +39/-4
  • Gender: Male
  • Air Chemistry Man
    • Supplier of air sampling equipment and services
Re: GC method using RRF instead of all compounds direct quantitation
« Reply #1 on: September 24, 2010, 03:43:03 PM »
Use of RRF is quite common and often is a standard feature in GC software. For better accuracy you will need to measure all RRF's for your target compounds. The approach works fine as long as calibration is linear when "forced through zero".

It does not work for nonlinear calibration curves:
-if detector is nonlinear (let say ECD)
-when you work very close to the detection limit;
-when concentration is too high and detector response is non-linear.
- Interferences 

Sponsored Links