GC-FID
« on: May 01, 2012, 12:28:28 AM »
I have two FID (flame ionization detector) jets with different exit diameters. One of the jet allows my capillary column to go right through it to the other end/tip while the other does not allow the capillary column to go through the end as the "tip" is narrower.
I am not familiar with the FID jet so which jet should be used?

Offline JGK

  • Chemist
  • Full Member
  • *
  • Posts: 738
  • Mole Snacks: +66/-19
  • Gender: Male
Re: GC-FID
« Reply #1 on: May 01, 2012, 03:08:22 PM »
What information does the instrument manual give you?

Experience is something you don't get until just after you need it.

Offline Stepan

  • Chemist
  • Full Member
  • *
  • Posts: 358
  • Mole Snacks: +39/-4
  • Gender: Male
  • Air Chemistry Man
    • Supplier of air sampling equipment and services
Re: GC-FID
« Reply #2 on: May 02, 2012, 07:53:39 PM »
The column shouldn't go through the jet. Try the one with narrow opening first. It should give better signal.