Hi Osmium,
Yes, the matrix of my standards are the same as the samples.
We used axial a few months ago, and figured out that Radial is less sensitive and showed better results than the axial. I figured out that the Radial is best to use when your concentrations are high, and Axial for low concentrations.
I will look into the Ag line that's less sensitive. We've tried only 3 so far, the one line dubbled each sample's value, where the other two were near to correct, but still about 50ppm to low.
What we also found is that on the ICP-OES, you get a small part called the "quick disconnect or release". This part is connected on the machine where the spray chamber are located. This small part had a argon gas leak, and caused the plasma to go brighter and dimmer. So we figured this might also be a huge problem, for the sample's results aren't constant. So today I've replaced that part (after 3 month's wait), and discovered that one part is too small. But luckily the guy will arrive next week Monday to insert an argon filter. So only after everything is been taken care of, I can go on with my experiment and see what the results will be. We figured that when everything is fixed, and no leakages appears, and the results are still inconsistant, then there are something wrong with the ICP-OES itself.
The way we determine this, is to analyse the same sample, but instead of triplicate readings, we do 30 at once. So it analyzes the same sample each second for 30 times. And in that results, there should not be a 50ppm difference. The sample is homogenous and in liquid form.
But thanks so much for your help Osmium. I'll try that Ag line and see what happens. Appreciate your reply!
Bambi