I completed an AAS lab. It is an analysis of lead by standard addtion. I had a 0ppm solution with absorbance 0.018, 2ppm with 0.030nm, and 3ppm with 0.036.
I had to plot the absorbance and ppm. The x intercept of the line will be the concentration of the diluted sample with no standard addition.
After plotting, the number for the X-intercept comes out to be -3.0ppm. Is it possible to have negative ppm in this case?