1. The retention time of a sample will be dependent on the GC conditions. The length of the column, the carrier gas flow rates, the stationary phase thickness, the temperature ramping profile, etc will affect WHEN your sample comes out. The sample retention time may vary by an hour and still be a valid measurement with good resolution if one lab uses a 100 m column and another lab uses a 10 m column. Before making any conclusions, gather this information for each lab and post it.
If both labs use exactly the same conditions (i.e. as specified by an ISO or ASTM method), the retention times should be similar (not exact but within 10 % of each other).
2. If you are referring the the mass spectra, and assuming you are using electron impact ionization source at 70 kV, then yes. The spectra should look nearly identical with minor variations for a particular compound. Peak matching software is commonly used to associate compilations of intensity and m/z ratio to known compounds because the variations become very subtle when trying to match similar compounds (i.e. 2-methylhexane and 3-methylhexane).