Strong UV signals are due to conjugation, which does not exist in the given molecule. As consequence, there are two options:
1). Either you take a UV spectrum (in the same solvent as used in HPLC, in not being aqueous) of the given compound and see if the strongest absorption band is strong enough for UV detection.
2). Or, you use a refractive index detector.