Hello, I hope I came to the right place, and was hoping someone here might be able to help me out!
We're having an issue at my Lab developing a robust method for analyzing various metals/alloys that we produce for trace metallic impurities via ICP-OES. Its been going on for some time, and I figured it was time to reach out beyond work to see if anyone has any ideas that could help us along.
We produce various precious metals and alloys(Au, Ag, AuSn, AgCu, AuNi etc), and test them for ~44 different metallic impurities. As it stands we have two methods we use, one for gold based alloys and one for basically anything else., but we have a big problem with wavelength selection/interferences and getting results to agree with our more accurate instruments (GDMS). We would also like to eliminate any user subjectivity from our techs.We use Thermo ICAP 6000 series instruments running iTeva software and Our method is roughly as follows.
Digest 0.2 g of our sample in 2 mL of Aqua Regia, and dilute to 10mL with DI water, these are run with a correction factor of 50.
which is run against a calibration of
Blank: 50:50 Aqua Regia:Water
Standard: 5ppm of each element we are analyzing for in 50:50 Aqua regia:water
all using 50 microliters of Yttrium as an internal standard
The method is not bad for analyzing pure metals such as Silver, but our more complicated alloys give us a lot of headaches with false positives and negatives, across various wavelengths.
I'm of the opinion that we need specific methods for each alloy group (ie: AuSn, CuAu, AgCu, AuNiPd), etc) because each will have unique interferences that must be dealt with. Some other ideas we've been looking at
-Digesting more sample to meet the intensity of our standard, and get rid of our 50x correction factor
-Matrix matching our blank and standard to our sample, (so if we are running CuAu, making our blank/standard in a CuAu solution of the same ratio as our sample, instead of just Aqua regia:water
-Utilizing method of standard addition for each sample instead of a normal blank/standard (very time consuming)
-splitting up our standard into multiple standards so the elements don't interfere with each other in the calibration
We've shied away from IECs due to how many elements we are dealing with, but it is still an option.
I know this is slightly vague, and ideally one of our three senior chemists should really be the ones figuring this out, not us lab techs, but I'll answer any questions as best I can. If anyone has had any experience running for trace metal impurities on these instruments, let me know!