December 22, 2024, 03:46:26 PM
Forum Rules: Read This Before Posting


Topic: Measuring pH of reverse osmosis's water  (Read 1391 times)

0 Members and 1 Guest are viewing this topic.

Offline Fakes

  • Regular Member
  • ***
  • Posts: 9
  • Mole Snacks: +0/-0
Measuring pH of reverse osmosis's water
« on: February 05, 2019, 04:14:31 AM »
I know that it should be acidic. But when i am measuring it in 2 different bench pH meters the result is 9 (in both) and when i measure it using a pH paper strip it is around 6. Any ideas?

Offline chenbeier

  • Sr. Member
  • *****
  • Posts: 1337
  • Mole Snacks: +102/-22
  • Gender: Male
Re: Measuring pH of reverse osmosis's water
« Reply #1 on: February 05, 2019, 05:45:38 AM »
A pH probe needs conductivity. DiI-water or reversd osmoses water cannot measured like this. You have to add a little bit NaCl or other neutral salt to the water and measure again.

Sponsored Links