Isophthalate peak in GCMS
Poll
I am getting a prominent peak for Isophthalate for any type of compund injected
GCMS
0 (0%)
peak
0 (0%)
Total Members Voted:
0
« previous
next »
Print
Pages: [
1
]
Go Down
Topic: Isophthalate peak in GCMS (Read 3473 times)
0 Members and 2 Guests are viewing this topic.
agro
New Member
Posts: 4
Mole Snacks: +0/-0
Isophthalate peak in GCMS
«
on:
May 09, 2009, 09:18:01 AM »
Hi,
I am getting one prominent peak for all of the samples, that account for diiso pthalate from NIST liberary match. Is it could be from piece of rubber septa from injection port?
How do I can get rid of the peak or remove from column? Please explain
Logged
marquis
Full Member
Posts: 441
Mole Snacks: +38/-3
Re: Isophthalate peak in GCMS
«
Reply #1 on:
June 30, 2009, 02:46:41 PM »
Ran across this one the hard way.
Check the gloves being used for sample prep. There are some opaque white gloves that are mixtures of polymers. One polymer used includes phthalates as a plasticizer. Any contact between the glove and the test solution can lead to contamination.
If you are using the purple gloves, you should be ok.
Logged