Hi there
There are several different types of reverse phases available in the market. These include: C18, C8, C4, C1, C-cyclehyxyl, C-phenyl, C-cyano and C-adamantyl (C-adamantyl is also called C-12).
C18 is by far and away the cheapest of these materials. The main reason for the low cost of C-18 silica gel is the fact that it is produced in the largest quantity by most manufacturers. When a material is produced in large quantities, the cost of this material is reduced.
C18 is often called the "traditional reverse phase matrix". Because it has the highest degree of hydrophobicity, C18 silica gel interacts with the widest range of compounds and the interactions are generally more pronounced. The reason why C-18 is more hydrophobic than the other reverse phases is because the length of the carbon chains are longer (i.e. C18 is longer than C8, and C8 is longer than C4 etc).
In my view, the reason why anybody would choose C8 instead of C18 is because they are looking for a reverse phase matrix that has a lower degree of hydrophobicity. Lower hydrophobicity means faster retention for non-polar compounds, therefore your non-polar compounds would move down the column more readily with C8 than with C18. The C8 sorbent would be particularly useful if your compounds streak on C18 silica gel.
To summarise:
C8 may be used instead of C18 when shorter retention times are required. C4 provides even less retention towards non-polar compounds than both C18 and C8
I don't think acidity (as in your previous message) had anything to do with this. Maybe the chemist was simply looking for a sorbent that would allow the compounds to move faster - hence their use of C8 instead of C18.
If you have any more specific questions I would be happy to help you out - msg me and I will give you my email address.
Best wishes
Athan