1. I disconnected the column at the back end, removing the detector from the system. Thus the flow was pump, injector, guard column, column, waste. The results were the same as before, pressure held steady at 60 bar, followed by a slow rise to 100 bar and a spike above the pressure limit in the 30% to 50% methanol range.
This will tell you virtually nothing as the pressure in HPLC is generated in the area between the pump and the column. The solvent pressure post column is negligible (pressures inexcess of 100 psi can crack detector flow cells)
2. I disconnected the column at the front end, removing the column and detector from the system. Thus the flow was pump, injector, guard column, waste. This also produced a similar result. The pressure held steady at 30 bar, followed by a slow rise to 100 bar and a spike above the max pressure limit in the 30% to 50% methanol range.
Was this repeatable? Were you using blank solvent injections or actual sample injections?
3. Next, I disconnected the guard column. Thus the flow was pump, injector, waste. No spike in pressure occurred. The pressure held around 6 bar and I was able to run through the entire method with no problems.
OK
4. Finally, I reconnected just the column. The flow in this test was pump, injector, column, waste. In this test, the pressure held at about 43 bar for approx 6 min before slowly rising to 100 bar and then spiking above the max pressure limit. To confirm this wasn't a one time occurrence, I ran the test 2 more times. Every time the spike would occur between 6.8-7.2 minutes or 89-91% methanol.
Was this repeatable? Were you using blank solvent injections or actual sample injections?
If the exercises performed were done using actual samples for the injections, I suspect that you have a precipitation issue. Something within the sample matrix is precipitating within either the guard or analytical column at high methanol concentrations. This precipitation is causing your blockage and pressure spike.