Atomic Absorption Test Method Help
« on: July 20, 2011, 04:08:50 PM »
I am trying to develop a method to test Titanium dioxide in creams and i keep constantly getting low numbers. I digest the sample with hydrofluoric acid/Nitric acid in a microwave digester, I get a clear colorless solution, with no percipitate. I adjust it to volume with water and then i run it in the AA, and every single time i get about 60-80% of the total concentration of Titanium dioxide... am i missing something?

Offline Stepan

  • Chemist
  • Full Member
  • *
  • Posts: 358
  • Mole Snacks: +39/-4
  • Gender: Male
  • Air Chemistry Man
    • Supplier of air sampling equipment and services
Re: Atomic Absorption Test Method Help
« Reply #1 on: July 20, 2011, 06:14:39 PM »
Digest the micro-sample (1-3 mg) with 1 mL HNO3 and 2 mL H2SO4 in a conical beaker and heat until about 1 mL remains. Quantitatively transfer the solution to a 25-mL volumetric flask, add 5 mL of 5,000 µg/mL potassium ion, then dilute to volume with DI H2O.

Use standards with the same concentration of potassium and sulfuric acid in DI water.

I did this test a month ago, recovery from TiO2 crystalline was more than 95%.

Offline blackpavvn

  • Very New Member
  • *
  • Posts: 2
  • Mole Snacks: +0/-0
Re: Atomic Absorption Test Method Help
« Reply #2 on: July 21, 2011, 07:55:21 AM »
Thank you, will try that