Retention Time
« on: November 04, 2011, 03:53:44 PM »
I have a sample that was run through GC/MS on two seperate occasions; the same instrument and the same analytical conditions were used on both occasions.

Both spectra appear identical. The only thing that is different is the retention time of some of the peaks. Of the 9 peaks present, the first 4 peaks have the same retention time in both spectra, whereas the retention time of the remaining 5 peaks are very slightly different, e.g. 3.70 and 3.71, 4.52 and 4.55...

Does anyone know why the last few peaks would have different retention times, even though its the same sample with the same equipment? Could it be something to do with using a manual injection prehaps?

Offline Stepan

  • Chemist
  • Full Member
  • *
  • Posts: 358
  • Mole Snacks: +39/-4
  • Gender: Male
  • Air Chemistry Man
    • Supplier of air sampling equipment and services
Re: Retention Time
« Reply #1 on: November 05, 2011, 04:21:42 PM »
Usually such shift is acceptable. It may occur in results of temperature variation in a GC oven.

Offline Toxophilus

  • New Member
  • **
  • Posts: 5
  • Mole Snacks: +1/-0
Re: Retention Time
« Reply #2 on: November 08, 2011, 07:33:29 AM »
I think your problem is thinking that they are spectra when in fact they are chromatograms. Spectral data is indeed very precise but retention times in any chromatographic technique will vary slightly; in GC this would be variation in both the gas flow and oven temperatures.

Offline TheUnfocusedOne

  • Full Member
  • ****
  • Posts: 103
  • Mole Snacks: +6/-5
  • Gender: Male
Re: Retention Time
« Reply #3 on: November 08, 2011, 02:12:38 PM »
+1 to this being normal. The later eluting compounds are more succeptible to variation since they've been in the column longer.

What's your method? Are you running isothermal or are you ramping?