You need to supply more information about your samples than the sample size, most importantly what analytes you are wishing to measure.
A general rule of thumb is to always use gas chromatography if it is possible, meaning the analytes to be measured in the samples are thermally stable and volatile (or can be derivatized to be made volatile such that they can be carried by the gas phase). The reason for this rule of thumb is the unsurpassed resolution and sharp peak shape in GC. Combined with the flame ionization detector, the GC-FID is very sensitive and low maintenance instrument for most organic compounds. Identity of compounds can be predicted based on using the Kovats retention index (based on the column you have -- go to
http://webbook.nist.gov/chemistry/) but true identification will be only possible using GC-MS (mass spec). If you know your sample though and have access to journal articles, this will show you what to expect for compounds and this makes it a lot easier. For sugars, organic acids and amino acids, MSTFA is the most powerful derivatizing agent and will make these compounds able to be analyzed by GC.
Otherwise, HPLC is your next step up. Here you need to find what will dissolve your compound (reverse phase for water/methanol/acetonitrile soluble compounds and normal phase for compounds soluble in more hydrophobic organic solvents). Also, you will need to select the right detector to use. Where there is conjugated dienes (such as phenolics, carotenoids or proteins due to tyrosine and tryptophan), a variable wavelngth or diode array detector may be used. Otherwise, an evaporative light scattering detector or refractive index detector (no gradient elution) may be required for things such as sugars or fats.
Finally, if you are working with large polymers, you will need to use size exclusion chromatography which can be classified as gel permeation chromatography (hydrophilic water soluble polymers) or gel filtration chromatography (hydrophobic organic soluble polymers), and then you will have to select your column based on the MW range of the compounds you wish to separate. Resolution with this technique is rather poor, so ofter it is just used to characterize your sample based on sMW size distribution rather than separating by individual compounds. If you have proteins, SDS-PAGE (electrophoresis) with quantification using densitometry will be far better for resolution of proteins.