i am using a chemilluminescense gas analyzer that measures NO, the analyzer measures bottled NO concetrations very well, when diluted entirely with N2. when i dilute 10% O2 into a NO/N2 sample, i find that the analyzer measures NO quite lower than the true NO concentration. FTIR sees NO2 forming, and its well known the chemilluminescense detector cant measure NO2. i read that the oxidation of NO into NO2 has a negative activation energy, such that cold temperatures will increase the oxidation. at ambient temperatures, am i seeing this reaction occur? i pass my sample through a 4 Celcius thermoelectric cooler to remove water, is it true this will cause even stronger, more rapid NO oxidation?