December 26, 2024, 12:40:52 PM
Forum Rules: Read This Before Posting


Topic: Column Chromatography Solvent to Separate 2 Nonpolar Compounds?  (Read 11498 times)

0 Members and 1 Guest are viewing this topic.

Offline yunie

  • Very New Member
  • *
  • Posts: 1
  • Mole Snacks: +0/-0
Column Chromatography Solvent to Separate 2 Nonpolar Compounds?
« on: October 24, 2012, 08:14:45 PM »
I have to separate two nonpolar compounds that are fairly close in polarity using column chromatography. I can't do reverse phase chromatography so the only thing I can really play around with is the solvent system. Should I try using a purely nonpolar solvent and then start using more polar mixtures or should I start with more polar mixtures and move to more nonpolar? I've already tried moving from an 80:20 Hexanes-Ethyl Acetate to 20:80 Ethyl Acetate:Hexanes, but that didn't separate them. Any help would be much appreciated!

Offline vex

  • Regular Member
  • ***
  • Posts: 66
  • Mole Snacks: +13/-0
  • Gender: Male
Re: Column Chromatography Solvent to Separate 2 Nonpolar Compounds?
« Reply #1 on: October 26, 2012, 01:42:10 PM »
You can try using TLC to test your systems before you run the big column. I would imagine that you won't be able to separate them with a very nonpolar solvent system, and that some intermediate mixture will be your best bet. You may also need a very long column if they're very similar.

Investigate different solvent systems using TLC, and if you are having trouble finding one that works, look into alternate purification methods, like recrystallization or sublimation.
University of Michigan Ph. D. Pre-Candidate, Inorganic Chemistry

Do or do not. There is no "try."

Offline Ligte

  • Regular Member
  • ***
  • Posts: 10
  • Mole Snacks: +2/-0
Re: Column Chromatography Solvent to Separate 2 Nonpolar Compounds?
« Reply #2 on: November 07, 2012, 01:57:35 PM »
You have to remember that the resolution or separation efficiency of your column is determined by your particle size. In your case I assume you are working at atmospheric pressure which means your gel has a large particle size otherwise there would be no flow through your column. In simple terms you will have great difficulty separating compounds with similar polarity the lower the resolution of your column and especially when you don’t use gradient elution.
What you can do
1.   Make the injection volume as low as possible and use smaller concentration.
2.   The polarity of the solvent or mixture of solvents uses to dissolve you compounds relative to the polarity of your mobile phase plays a major role in separation.
3.   When using TLC plates what you can try is to use for example the (8:2) Hexane-Ethyl Acetate solution to run the first separation then after you dried your plate, turn that same TLC plate on its side and then do the run again with a (2:8) Hexane-Ethyl Acetate solution otherwise just use gradient elution.
Using the same low resolution packing material and increasing its column length with the same solvent system is time consuming, expensive and bad science.

Offline OC pro

  • Chemist
  • Full Member
  • *
  • Posts: 396
  • Mole Snacks: +36/-15
  • Gender: Male
Re: Column Chromatography Solvent to Separate 2 Nonpolar Compounds?
« Reply #3 on: November 07, 2012, 03:19:40 PM »
For very nonpolar compounds, ethyl acetate is often too polar. Means it will push everything through silica gel. Try to use diethyl ether or tert butyl methyl ether in combination with hexane. Start with a 1:1 mixture. I accomplished even separation using petrol ether/diethyl ether 100:1 mixtures.

Offline Babcock_Hall

  • Chemist
  • Sr. Member
  • *
  • Posts: 5715
  • Mole Snacks: +331/-24
Re: Column Chromatography Solvent to Separate 2 Nonpolar Compounds?
« Reply #4 on: November 07, 2012, 07:45:51 PM »
I have separated carotenoids (very nonpolar ones, such as spirilloxanthin) on silica with 15% acetone in hexanes.  When I separated them on alumina, I recall using even less acetone (perhaps less than 10%), but my memory is imperfect.  Are you using TLC plates to try different solvent systems first, before running a column?

Offline Dan

  • Retired Staff
  • Sr. Member
  • *
  • Posts: 4716
  • Mole Snacks: +469/-72
  • Gender: Male
  • Organic Chemist
    • My research
Re: Column Chromatography Solvent to Separate 2 Nonpolar Compounds?
« Reply #5 on: November 08, 2012, 04:30:00 AM »
My research: Google Scholar and Researchgate

Sponsored Links