OMNIC software - background before/after
« on: February 16, 2013, 08:13:49 PM »
Hey guys,

Quick question I was using a Nicolet FTIR with OMNIC 8.2 software installed this week and I noticed when I first collected the background spectrum (in this case, water) it asked if it wanted the software to collect the background 'before' or 'after' the sample spectrum.

Does this make a difference?

In the end the water background was subtracted.

Real world logic tells me I should of selected 'after' (I chose 'before') - hope my data isnt worthless now.

Offline Stepan

  • Chemist
  • Full Member
  • *
  • Posts: 358
  • Mole Snacks: +39/-4
  • Gender: Male
  • Air Chemistry Man
    • Supplier of air sampling equipment and services
Re: OMNIC software - background before/after
« Reply #1 on: February 18, 2013, 02:09:23 AM »
I was under impression that background scan assumes a blank scan without a cuvette (just air). Usually it is taken first before any measurements and often used by the instrument for automatic self-tuning and diagnostic.

Offline lespaul

  • Regular Member
  • ***
  • Posts: 50
  • Mole Snacks: +0/-1
Re: OMNIC software - background before/after
« Reply #2 on: March 17, 2013, 12:27:05 PM »
Thanks Stephan,

Well it looks like the data is okay, I guess it doesnt really matter if you select 'before' or 'after' when subtracting a background (ie. water, air, solvent, etc).

Im new to OMNIC