The TIC I showed are acquiring only acquiring data from 500-2000m/z, because that is where my analytes will be. Is that reason I am not seeing the noise that you were talking about? If so, to what value should I set those parameters?
Don't change your SCAN acquisition parameters (m/z 500-2000 in X sec), but do adjust the DATA processing parameters (minimum peak width, minimum peak area, and one other that I cannot remember (Peak Threshold ??)) to acquire low-level noise. Try setting these while you scan m/z 150-500 in X secs-----N.B. do note the data processing params. before changing---it may only be the peak threshold that needs to be lowered to pick up the required noise.
Did you mean to inject the sample three times without chromatography?
No, I did mean with chromatography; however, to save time and to observe the base-line, you could initially do three shots without a column.
My MeCN/H20 ratio is about 30:70. That is the mobile phase composition at which the compound elutes from the LC column. I am considering adding 70:30 MeCN:H20 w/ 0.1% formic acid post-column to try to increase the organic content of the mobile phase and enhance ionization. Do you think that will give me a more consistent signal?
Since ESI signal is concn.-dependent I would not add a post-column diluent, UNLESS sensitivity is a real issue and the increased organic content may overcome the dilution effect. In your case here, I would change from MeCN/H2O to MeOH/H2O as a first step; methanol is a weaker eluent, and you will probably need 40:60 rather than 30:70 to achieve the same R.T. This should improve sensitivity considerably, and 0.1% formic acid would not hurt.
Keep us posted.