Hi could someone please explain how to calculate the following,

ICP-OES gives a sample's concentration reading of 29000 mg/kg (after acid digestion).
I used 0.3 g of sample and diluted with deionised water to a final volume of 50ml.
I would like to know the actual concentration. (mg/kg).

Thank you.

Offline Corribus

  • Chemist
  • Sr. Member
  • *
  • Posts: 3551
  • Mole Snacks: +546/-23
  • Gender: Male
  • A lover of spectroscopy and chocolate.
Re: ICP OES dilution calculation
« Reply #1 on: June 19, 2016, 11:43:07 AM »
Your measured concentration value just needs to be multiplied by the assayed volume to give a total amount of analyte in the sample. Then divide this by the sample mass to get the mass of analyte per mass unit of sample.
What men are poets who can speak of Jupiter if he were like a man, but if he is an immense spinning sphere of methane and ammonia must be silent?  - Richard P. Feynman

Offline jvm208

  • Regular Member
  • ***
  • Posts: 17
  • Mole Snacks: +0/-0
Re: ICP OES dilution calculation
« Reply #2 on: June 27, 2016, 07:31:44 AM »
Thank you so much  :)